ISIS - effect of tilt on wavelength calibration

Design, construction, tuning of spectroscopes
Information and discussion about softwares (telescope remote, autoguiding, acquisition, spectral processing ...)
Post Reply
Robin Leadbeater
Posts: 1957
Joined: Mon Sep 26, 2011 4:41 pm
Contact:

ISIS - effect of tilt on wavelength calibration

Post by Robin Leadbeater »

In IRIS the tilt algorithm uses a projection onto the X axis rather than a rotation algorithm. This gives fewer artifacts but changes the dispersion by cos(tilt) Is the same algorithm used in ISIS?

Thanks
Robin
LHIRES III #29 ATIK314 ALPY 600/200 ATIK428 Star Analyser 100/200 C11 EQ6
http://www.threehillsobservatory.co.uk
Christian Buil
Posts: 1424
Joined: Mon Sep 26, 2011 6:59 pm
Contact:

Re: ISIS - effect of tilt on wavelength calibration

Post by Christian Buil »

Yes Robin, ISIS use X-projection for correct the tilt, like IRIS. But note an important point : The calibration emission lines are also rotated in a similar way, so "cos" calibration deviation is not a problem (of course, it is necessary to have the same geometry for science images and calibration images - the basic for good calibration).

Christian
Robin Leadbeater
Posts: 1957
Joined: Mon Sep 26, 2011 4:41 pm
Contact:

Re: ISIS - effect of tilt on wavelength calibration

Post by Robin Leadbeater »

Thanks Christian,

It is important for Star Analyser users to understand that if they use ISIS, they should not use a dispersion equation calculated for a different angle. They will not expect this if they normally use the rotation function in VSpec or RSpec where the dispersion does not change with angle.

Cheers
Robin
LHIRES III #29 ATIK314 ALPY 600/200 ATIK428 Star Analyser 100/200 C11 EQ6
http://www.threehillsobservatory.co.uk
Post Reply