Re: evaluation of noise and gain for QHY6Pro
Posted: Tue Jun 26, 2012 10:49 pm
Hello Dong Li,
I guess Maxim has a somewhat special display of the histogram with a logarithmic scale for the vertical axis. A Gaussian has the following shape, taken from a bias of my camera (SBIG ST8300): The logarithmic scale changes the Gaussian shape into a characteristic sugar loaf shape.
The logarithmic scale you can see if you open process - levels and stretch the intensity values so you can see the principal intensity range: In this display the vertical scale is plotted, down to 1, so you see also rare warm or hot pixels. The pedestal is cut off in the previous screen stretch histogram.
The pedestal in your histogram indicates that apart from the gaussian noise you have some pixels with a wider distribution of intensities. This might indicate some additional noise from readout, I have seen this with poor quality cables, if the cable picks up some noise or ripple on the power supply. With a linear scale this hardly shows up so you can work with your CCD.
If the pedestal comes from a constant nonuniformity of the pixels it might be subtracted out during processing. For measuring readout noise you take the difference between two bias images. If you have some fixed pattern noise the diffference bias should show lower noise, if the noise is completely random, the difference noise is larger by a factor sqare root 2.
Unfortunately the documentation of Maxim does not say exactly what is displayed in the histogram, you normally don't bother about it for making nice images.
Do not use Histogram - levels, this destroys the linearity of your camera response and makes the images useless!
I hope this clarifies somewhat what you measure. It is a good idea to check your camera before using it, readout noise, dark current and linearity are essential for good quality spectra.
Regards, Martin
I guess Maxim has a somewhat special display of the histogram with a logarithmic scale for the vertical axis. A Gaussian has the following shape, taken from a bias of my camera (SBIG ST8300): The logarithmic scale changes the Gaussian shape into a characteristic sugar loaf shape.
The logarithmic scale you can see if you open process - levels and stretch the intensity values so you can see the principal intensity range: In this display the vertical scale is plotted, down to 1, so you see also rare warm or hot pixels. The pedestal is cut off in the previous screen stretch histogram.
The pedestal in your histogram indicates that apart from the gaussian noise you have some pixels with a wider distribution of intensities. This might indicate some additional noise from readout, I have seen this with poor quality cables, if the cable picks up some noise or ripple on the power supply. With a linear scale this hardly shows up so you can work with your CCD.
If the pedestal comes from a constant nonuniformity of the pixels it might be subtracted out during processing. For measuring readout noise you take the difference between two bias images. If you have some fixed pattern noise the diffference bias should show lower noise, if the noise is completely random, the difference noise is larger by a factor sqare root 2.
Unfortunately the documentation of Maxim does not say exactly what is displayed in the histogram, you normally don't bother about it for making nice images.
Do not use Histogram - levels, this destroys the linearity of your camera response and makes the images useless!
I hope this clarifies somewhat what you measure. It is a good idea to check your camera before using it, readout noise, dark current and linearity are essential for good quality spectra.
Regards, Martin