Hello,
Just wanted to show an other simple method in order to calibrate low resolution spectra with the Neon lmap available with LISA and LHIRES3.
- First of all, take a short (typically 0.3 s) exposure spectrum of the lamp in order to get red Neon lines;
- Take a long exposure (typically 15 s, be aware with remanance!) to get blue lines;
- Merged 2d spectrum with the background unified (red edge of long exposure was set to 0, see Juanary SpcAudace release):
You get the best of the two worlds!
Then, the lines profiles with identified main lines is made for you:
(Figure updated on 2012/01/29)
I recommand you to select widespreaded lines in order not to favour the red edge or blue edge.
- The following lines seem working fine for a deg 3 polynom:
4065.65, 4348.06, 5400.56, 5852.49, 6402.25, 7032.41 and 7245.17
Which lead to a rms of 0.6 A where the dispersion is here 2.57 A/pixel, so an accuracy of 1/5 pixel.
- With a deg 4 polynome and the following lines:
4065.65, 4348.06, 4710.06, 5400.56, 5852.49, 6402.25, 7032.41, 7245.17, 6143.06, 6678.28 and 6929.47,
the following dispersion law was found: Lambda=3917.50781533+2.43491203042*(x-1)+0.000230671871345*(x-1)^2+-1.61973682859e-07*(x-1)^3+4.63971559225e-11*(x-1)^4 with RMS=0.293443870835,
So a linear law described by: 3917.50781533+2.56720166243*(x-1).
Enjoy,
Benji
Just wanted to show an other simple method in order to calibrate low resolution spectra with the Neon lmap available with LISA and LHIRES3.
- First of all, take a short (typically 0.3 s) exposure spectrum of the lamp in order to get red Neon lines;
- Take a long exposure (typically 15 s, be aware with remanance!) to get blue lines;
- Merged 2d spectrum with the background unified (red edge of long exposure was set to 0, see Juanary SpcAudace release):
You get the best of the two worlds!
Then, the lines profiles with identified main lines is made for you:
(Figure updated on 2012/01/29)
I recommand you to select widespreaded lines in order not to favour the red edge or blue edge.
- The following lines seem working fine for a deg 3 polynom:
4065.65, 4348.06, 5400.56, 5852.49, 6402.25, 7032.41 and 7245.17
Which lead to a rms of 0.6 A where the dispersion is here 2.57 A/pixel, so an accuracy of 1/5 pixel.
- With a deg 4 polynome and the following lines:
4065.65, 4348.06, 4710.06, 5400.56, 5852.49, 6402.25, 7032.41, 7245.17, 6143.06, 6678.28 and 6929.47,
the following dispersion law was found: Lambda=3917.50781533+2.43491203042*(x-1)+0.000230671871345*(x-1)^2+-1.61973682859e-07*(x-1)^3+4.63971559225e-11*(x-1)^4 with RMS=0.293443870835,
So a linear law described by: 3917.50781533+2.56720166243*(x-1).
Enjoy,
Benji
Last edited by Benjamin Mauclaire on Sun Jan 29, 2012 3:47 pm, edited 5 times in total.
Spcaudace spectroscopy software: saving you hundred hours of frustration.
-
- Posts: 135
- Joined: Thu Sep 29, 2011 9:36 pm
Re: Low resolution calibration method
Hi Benji,
Thanks a lot for that simple way for getting the blue side.
Will try it next time I am in the Observatory
Thanks a lot for that simple way for getting the blue side.
Will try it next time I am in the Observatory
regards Rainer____Observatorio Real de 14_____MÉXICO_____N 22° W 101°
-
- Posts: 14
- Joined: Fri Sep 30, 2011 3:25 pm
Re: Low resolution calibration method
Nice work Benjamin. I hope repeated long neon exposures are not harmful for the acquisition camera... Also I am afraid we still lack calibration wavelengths between 4348 and 5341Angstroems. This may hamper the accuracy of a high degree (>3) calibration polynomial.
Patrick Lailly
Patrick Lailly
-
- Posts: 258
- Joined: Thu Sep 29, 2011 10:14 am
Re: Low resolution calibration method
Patrick,
You're right, with a deg 4 polynom, you may also ty using 4710 blended Neon line.
I improved calibration by finding that blue lines were blended.
Note the updated calibrated lines profile.
Benji
You're right, with a deg 4 polynom, you may also ty using 4710 blended Neon line.
I improved calibration by finding that blue lines were blended.
Note the updated calibrated lines profile.
Benji
Spcaudace spectroscopy software: saving you hundred hours of frustration.
-
- Posts: 14
- Joined: Fri Sep 30, 2011 3:25 pm
Re: Low resolution calibration method
I tried to calibrate my low resolution spectra using these lines but the result shows an inaccuracy of about 5 Angstroems in the blue. Did someone check that the neon lamp that equips the Lisa shows the mentioned wavelengths?Benji wrote:Hello,
- With a deg 4 polynome and the following lines:
4065.65, 4348.06, 4710.06, 5400.56, 5852.49, 6402.25, 7032.41, 7245.17, 6143.06, 6678.28 and 6929.47,
Benji
Patrick Lailly
-
- Posts: 14
- Joined: Fri Sep 30, 2011 3:25 pm
Re: Low resolution calibration method
After carrying out a detailed analysis of the lines shown by the neon calibration lamp between 3900 and 4300 A, I made extensive experiments to reduce the inaccuracy mentioned above and I conclude that the attractive approach suggested by Benji seems hopeless. We cannot avoid using the Balmer lines of a reference star for an accurate calibration.
Patrick Lailly
Patrick Lailly
-
- Posts: 135
- Joined: Thu Sep 29, 2011 9:36 pm
Re: Low resolution calibration method
Hi,
Finally I was able to test the calibration proposed by Benji.
I made 10 Neon images with 5 seconds and 10 images with 300s and then stitched them together. Then I analized it and the result is quite good from my point of view as I compared with an equation I use made on Vega.
If I have the opportunity I will test the calibration method on various stars.
Below you can see both results.
Integrated Neon calibration lamp in my LISA spectrograph with 35µm slit.
Vega calibration
Finally I was able to test the calibration proposed by Benji.
I made 10 Neon images with 5 seconds and 10 images with 300s and then stitched them together. Then I analized it and the result is quite good from my point of view as I compared with an equation I use made on Vega.
If I have the opportunity I will test the calibration method on various stars.
Below you can see both results.
Integrated Neon calibration lamp in my LISA spectrograph with 35µm slit.
Vega calibration
regards Rainer____Observatorio Real de 14_____MÉXICO_____N 22° W 101°
-
- Posts: 1521
- Joined: Fri Sep 23, 2011 1:01 pm
- Location: Rouen
- Contact:
Re: Low resolution calibration method
Hi Rainer,
The resolution of your spectrum looks very good.
Which slit did you use ?
Could you send us the .dat file ?
All the best
François Teyssier
The resolution of your spectrum looks very good.
Which slit did you use ?
Could you send us the .dat file ?
All the best
François Teyssier
François Teyssier
http://www.astronomie-amateur.fr
http://www.astronomie-amateur.fr
-
- Posts: 135
- Joined: Thu Sep 29, 2011 9:36 pm
Re: Low resolution calibration method
Hi Francois,
Which spectrum do you refer to ? Combined Neon ? If yes please download from this link
http://astroforo.net/astro/rspec/LISA/N ... s02762.dat
and here the combined Neon image *.fit file
http://astroforo.net/astro/rspec/LISA/N ... s-0300.tif
I used the 35µm slit.
regards Rainer
Which spectrum do you refer to ? Combined Neon ? If yes please download from this link
http://astroforo.net/astro/rspec/LISA/N ... s02762.dat
and here the combined Neon image *.fit file
http://astroforo.net/astro/rspec/LISA/N ... s-0300.tif
I used the 35µm slit.
regards Rainer
regards Rainer____Observatorio Real de 14_____MÉXICO_____N 22° W 101°
-
- Posts: 1521
- Joined: Fri Sep 23, 2011 1:01 pm
- Location: Rouen
- Contact:
Re: Low resolution calibration method
Hi Rainer,
I think there's a problem.
Your lines don't match with the theorical ArI lines given by VSpec.
I took a 120 sec spectrum of the neon lamp. Divided by its "continuum". Wavelenght calibrated (ISIS)
Your spectrum : black
Mine : blue
ArI : orange
NeI : cyan
Neon lines near 5400 match correctly
I'am taking another spectrum, 600 secondes exposure in order to get the weak lines.
I'll keep you informed
All the best
François
I think there's a problem.
Your lines don't match with the theorical ArI lines given by VSpec.
I took a 120 sec spectrum of the neon lamp. Divided by its "continuum". Wavelenght calibrated (ISIS)
Your spectrum : black
Mine : blue
ArI : orange
NeI : cyan
Neon lines near 5400 match correctly
I'am taking another spectrum, 600 secondes exposure in order to get the weak lines.
I'll keep you informed
All the best
François